There are several meters out there, but pinpoint seems to be one of the most popular ones. Check out some of the sponsors of RDO and see if any of them offers good deal.
As far as the accuracy/precision, I am going to say that there is little difference between 1% and 2%. There are several things that I want to you be aware of:
1, most accuracy is full scale deviation, not the reading that you see. So if you see 10 uS, and the accuracy is ± 1 %, is is not 10±0.1 uS, but rather, 10±(1999uS*0.01) or 10±19uS. You really need to read the specs carefully.
2, most TDS meter will give you only one digit of readout when you use it on RO water. My RO reading is 13ppm, for example. For RO/DI, almost all TDS meter is useless when it comes to monitoring the DI resin, since all you see is "0" ppm (underrange reading, not that it is perfectly pure).
3, fortunately, most of the hobbyists were able to make good use of TDS/conductivity meter despite the short comings mentioned above, but I just want people to be aware of them.